There appears to be a conflict between 660 and 517 regarding the sizing of an X-Ray feeder sizes. I'm wondering if I'm missing some nuance.
660.6(B): indicates that the rated ampacity of conductors and overcurrent devices fo a feeder for two or more branch circuits shall be 2 times the momenatary of the two largest equipment plus 20% of the momentary of all other machines.
517.73 Rating of Supply Conductors and Overcurrent Protection, (A)(2) indicates that feeders and overcurrent protective devices shall be 50% of the momentary for the lartest, plus 25% of the momentary for each one after that plus 10% of the momentary of the rest of the machines.
Do you agree that these are in direct conflict?
Thanks,
Mike
660.6(B): indicates that the rated ampacity of conductors and overcurrent devices fo a feeder for two or more branch circuits shall be 2 times the momenatary of the two largest equipment plus 20% of the momentary of all other machines.
517.73 Rating of Supply Conductors and Overcurrent Protection, (A)(2) indicates that feeders and overcurrent protective devices shall be 50% of the momentary for the lartest, plus 25% of the momentary for each one after that plus 10% of the momentary of the rest of the machines.
Do you agree that these are in direct conflict?