Help!
I just finished a project for an urgent care office that had a x-ray room, not MRI, an this is some of the things we did. You need to work hand in hand with the manufacturer (We had Siemens X-Ray) of the equipment to locate, j-boxes, disconnectes, epo switches, underground conduit to table. We did no wiring of the equipment just the power to disconnect and to j-box for equipment. We also pulled wiring for epo switch. Another thing that was used in this room was lead backed sheetrock, everything needed to be in the walls and double checked by rep from the manufacturer before it was inclosed, hard to cut the lead-sheetrock. Get a good set of plans from the manufacturer.