I have a requirement of do an A/C hi-pot test with a minimum of 500VA for a CE marking. We hired a compliance consultancy agency to do the first run of our equipment, and he advised us that we do not have to do the AC hi-pot test because our equipment has noise suppression devices and could be damaged from the AC hi pot test. He instead proceeded to do a DC hi pot test on our equipment.
We do have quincharcs across our solenoids and contactor coils, would this be what he meant by noise suppression? Does the compliance engineer's advice seem valid?
61010 (the IEC standard I am using) allows room to not do prescribed tests if you have a valid reason for not doing the test. For example, if you can show through manufactures information that the equipment would undoubtedly pass the test, then you do not have to perform the test.
Reasoning from this, since the AC test could damage the equipment then a DC hi pot test can be substituted. I am trying to figure out how to equate the 2 tests. For example, how can I say that a DC test performed for 1 minute an 2100V is equivalent to an 1 minute AC test of 1500V.
We do have quincharcs across our solenoids and contactor coils, would this be what he meant by noise suppression? Does the compliance engineer's advice seem valid?
61010 (the IEC standard I am using) allows room to not do prescribed tests if you have a valid reason for not doing the test. For example, if you can show through manufactures information that the equipment would undoubtedly pass the test, then you do not have to perform the test.
Reasoning from this, since the AC test could damage the equipment then a DC hi pot test can be substituted. I am trying to figure out how to equate the 2 tests. For example, how can I say that a DC test performed for 1 minute an 2100V is equivalent to an 1 minute AC test of 1500V.