Why is there a 208V rating for this X-Ray generator if it can't operate at 208V...? :?
I must have missed something.. It says use an auxiliary boost transformer if line voltage is 208 or less.
I dont see a 208 rating anywhere, so a buck-boost is indeed required.
They aux. boost transformer is an option usually provided by the xray manufacturer or the xray vendor/installer.
So if they provide the transformer, you know how much current it will draw, and what size breaker and wire to provide, etc.