I opened the case of my AT2004 today and noticed that one of the cables had broken. Thought I would start here to see where and how I can get a replacement. See images below

Thanks in advance.
 

glene77is

Senior Member
Location
Memphis, TN
... AT2004 ... one of the cables had broken.

Check with www.LicensedElectrician.com.
I have had excellent service and advice from them, for my At-4001 set.

I think you are looking for the Amprobe C2901 cordset.

Locally I have bought suitable cables with the covered plug-ends from Johnstone (HVAC supplier) since they handle multi-meters.
I have also used our local Electronic supply house for special end connectors
(sometimes depends on who you know).

I have also bought extra cables and cut them in the middle,
then re-soldered them to make male-male and female-female cords,
with shrink tubing over the joint.
Something with a 600V rating !
 

iwire

Moderator
Staff member
Location
Massachusetts
Seems to be a common issue with those leads, I soldered mine a couple times.

After that I gave up and just use the leads from one of my DMMs.
 

glene77is

Senior Member
Location
Memphis, TN
I'll have to give that a try next time.

17,
Well, that gave me a tingly feeling.
I was using the AT-4001 signal tracer,
had one of the transmitter leads strung out 25' to an outside water line,
had a live signal going out.

When I bumped the cable,
the alligator clip popped off and drew sparks. :confused:
This unit works very well, but I was surprised that it drew sparks.
This sort of puzzle interests me.
Now I am curious about what kind of circuitry is inside.

My DataComm signal tracer system never did that,
or at least I have not set it up this way and seen this activity.
 

iwire

Moderator
Staff member
Location
Massachusetts
My understanding is that the Amprobe unit places a pulsing current on the circuit conductors. That is why you can only trace the signal between the sender unit and the source.
 

glene77is

Senior Member
Location
Memphis, TN
My understanding is that the Amprobe unit places a pulsing current on the circuit conductors. That is why you can only trace the signal between the sender unit and the source.

iWire,
Sounds about right to me.
On a hot circuit,
this would be a Current Pulse mode of tracing.

In tracing a dead circuit,
I connect one side of the transmitter into the black wire,
and run the other transmitter line out to a water pipe ground.
This time, the signal ticks can be traced everywhere on that wire.
This would be a Voltage Pulse mode of tracing.

So, now I need to experiment on tracing a hot circuit,
going away from the AC source.

Will the fun never stop!
 