Recently had numerous outlets in a medical procedure area where plug tester was showing open ground. I had been asked to install gfi protected outlets as this is considered a "wet" area. Medical equip tech informed me that there is a "floating ground system" present. What is this? How does it work? Does it negate the need for GFCI protection of personnel?
 

roger

Moderator
Staff member
Location
Fl
Occupation
Retired Electrician
Sounds like an Isolated Power System, see 517.160. Is there a panelboard with a Line Isolation Monitor in the area?
If it is an Isolated Power System your plug tester will show faults, the system does not have a neutral, it is 60v / 60v L-L. You do not need GFCI's on these systems.

Roger
 

kwired

Electron manager
Location
NE Nebraska
Sounds like an Isolated Power System, see 517.160. Is there a panelboard with a Line Isolation Monitor in the area?
If it is an Isolated Power System your plug tester will show faults, the system does not have a neutral, it is 60v / 60v L-L. You do not need GFCI's on these systems.

Roger
Yes, make sure it isn't one of these before proceeding with anything. GFCI could even be undesired here, there is a reason it is set up like it is, if that is what it is.
 