I would like to get some opinions on an issue I'm coming across right now.
I'm working on a women's health clinic that has two Digital Mammo x-ray machines in the space. Currently I have a 480/277V 3 phase with a 45KVA
step down transformer feeding a 208/120V 3 phase panel.
Currently there is a 2.4KW 1 phase a/c unit on the same 208/120V 3 phase panel. The mammo x-ray equipment vendor is insisting that a power conditioner be put on the dedicated branch circuit feeder each of the mammo x-ray machines. The mammo x-ray equipment manufacturers requirements call for a maximum line impedance not to exceed 0.20 ohms for a 208V line.
Does anyone think that by having the a/c unit on the same panel will affect the mammo x-ray equipment and do you think it would exceed the 0.20 ohms requirement just by having the a/c unit feeding off the same panelboard?
Any input or suggestions would be greatly appreciated. Thanks.
I'm working on a women's health clinic that has two Digital Mammo x-ray machines in the space. Currently I have a 480/277V 3 phase with a 45KVA
step down transformer feeding a 208/120V 3 phase panel.
Currently there is a 2.4KW 1 phase a/c unit on the same 208/120V 3 phase panel. The mammo x-ray equipment vendor is insisting that a power conditioner be put on the dedicated branch circuit feeder each of the mammo x-ray machines. The mammo x-ray equipment manufacturers requirements call for a maximum line impedance not to exceed 0.20 ohms for a 208V line.
Does anyone think that by having the a/c unit on the same panel will affect the mammo x-ray equipment and do you think it would exceed the 0.20 ohms requirement just by having the a/c unit feeding off the same panelboard?
Any input or suggestions would be greatly appreciated. Thanks.