We have bring 20A circuit, protected by 20A/1P breaker to sterilizer with information as indicated below from equipment cutsheet:
Dedicated line 120V, NEMA 5-20R outlet, 20A circuit, wire gaurd sufficient to maintain 16A continuous load with a 40A momentary current.
They have told us when equipment started, the CB tripped. Do you know what percentage can 120V - 20A/1PH CB can be overloaded? I think momentary current is not a factor in this case and not sure what CB is tripped while manufacturer cutsheet indicates we only need to provide only
20A circuit as stated above, even they did not state about CB size requirement but we can go by outlet and circuit size.
Dedicated line 120V, NEMA 5-20R outlet, 20A circuit, wire gaurd sufficient to maintain 16A continuous load with a 40A momentary current.
They have told us when equipment started, the CB tripped. Do you know what percentage can 120V - 20A/1PH CB can be overloaded? I think momentary current is not a factor in this case and not sure what CB is tripped while manufacturer cutsheet indicates we only need to provide only
20A circuit as stated above, even they did not state about CB size requirement but we can go by outlet and circuit size.