517.72(A) refers to a required disconnecting means for a medical xray installation. The Xray company typically wants this disconnect installed right by the xray generator so their service tech has convinent access to the disconnect. Then they usually require an EPO switch at the operators station (behind the lead screen wall) for use by the operator.
517.72(B) says:
My question is what does "Xray control" mean? Do you take that to mean the operators location behind the lead screen wall? Our electrical inspector interpert's it that way, and I always have to show a second disconnect at the operators station.
Or could "Xray control" be similar to a "motor control" location? If this were the case, I would consider the xray generator to be the "control" location. And the single disconnect that the Xray vendor's drawings always show would suffice.
Steve
517.72(B) says:
Code:
Location. The disconnecting means shall be operable
from a location readily accessible from the Xray control.
Steve