Relocatable Power Taps
Microwatt
North Dakota
The UL white books states that relocatable power taps are not intended for use in patient care areas. There is a mfg. with a relocatable power tap on the market and they advertise this product can be used in a patient care area because it is UL 60601-1 and UL 60590-1 compliant. I guess my question is can this be the case? If the white book says they are not allowed can these other standards over-turn that?
Just a guess here but does that UL category say something like 're-locatable power taps for use in ordinary locations'?
Perhaps there is another category for 're-locatable power taps for use in patient care areas'?
Thanks iwire. I'm finding out that the ones listed for use in a patient care area are ones that come with a medical cart that itself is listed as an assembly. The power strips themselves are not listed for that application. Below is a clipping from a UL article.
Some UL Classified carts include
relocatable power taps as an integral
part of the cart. These are Recognized
Component Special Purpose Relocatable
Power Taps (SPRPTs), which are specifically
investigated to supply power to the plugconnected
equipment on the cart and to
verify compliance with the Standard for
Health Care Facilities, NFPA 99, Section
8.4.1.2.4.2. This standard is applicable when
two or more power receptacles supplied by
a flexible cord are used to supply
power to plug-connected components
of a movable equipment assembly that
is rack-, table- or pedestal-mounted.
SPRPTs have not been investigated for
stand-alone use in patient care vicinities.