    Dental Office

    Im thinking, would it be possible to run a homerun cable 12/3, 12/3, 12/3, 12/2 to a general location with regular MC and from there branch off with my hospital grade mc?

    a reel of the home run runs me about $1200 plus tax
    and a reel of the hospital grade runs me about $627 plus tax.
    Any ideas?

    Also any idea on which type of floor box would be best to install under a dental chair.
    Not familiar with any brands

  2. #2
    Join Date
    Feb 2003
    Location
    Fl
    Posts
    17,754
    Mentioned
    1 Post(s)
    Tagged
    0 Thread(s)
    517.13(A) says "All branch circuits serving" not "Part of all branch circuits serving". The branch circuit originates at a panelboard so that is where requirements of the section start. as far as the floor box, look into Steel City, Hubbell, or Walker.


    Roger
    Moderator

  3. #3
    Join Date
    Jun 2006
    Location
    Westchester County NY
    Posts
    272
    Mentioned
    0 Post(s)
    Tagged
    0 Thread(s)
    Not unless the home run MC cable sheath complies with 250.118

    See 517.13(A) also Article 100 definition of branch circuit.

    Also of note do forget 210.4(B) in regards to your multi wire branch circuits. (Disconnecting means)

    Would have to see the chair specs

    As for floor boxes i'm partial to Hubbell S1PFB as the low voltage and line voltage conduits are parallel in the box, so piping and trenching is easier, plus the low voltage entry is large.

    As for a poke thru take your pick.

  4. #4
    Join Date
    Jun 2006
    Location
    Westchester County NY
    Posts
    272
    Mentioned
    0 Post(s)
    Tagged
    0 Thread(s)
    Quote Originally Posted by Joe Villani View Post
    Not unless the home run MC cable sheath complies with 250.118

    See 517.13(A) also Article 100 definition of branch circuit.

    Also of note do forget 210.4(B) in regards to your multi wire branch circuits. (Disconnecting means)

    Would have to see the chair specs

    As for floor boxes i'm partial to Hubbell S1PFB as the low voltage and line voltage conduits are parallel in the box, so piping and trenching is easier, plus the low voltage entry is large.

    As for a poke thru take your pick.
    My mistake, the box I quoted won't work in a dental. Chair application.

    Sorry

  5. #5
    Join Date
    Aug 2006
    Location
    Miami Fla.
    Posts
    7,513
    Mentioned
    0 Post(s)
    Tagged
    0 Thread(s)
    All the chairs I have done the box was supplied by the equipment manufacturer.

  6. #6
    Join Date
    Apr 2006
    Location
    Simi Valley, CA
    Posts
    8,615
    Mentioned
    0 Post(s)
    Tagged
    0 Thread(s)
    Most of the chairs just plug in now days.
    I can build anything you want if you draw a picture of it on the back of a big enough check.

    There's no substitute for hard work....but that doesn't mean I'm going to give up trying to find one.

    John Childress
    Electrical Inspector
    IAEI / CEI / C10
    Certified Electrical Inspector

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •