220306-1207 EST
Since Larry is the only one to respond so far I will provide the answers. The purpose of the question is to provide a learning experience.
But first I would like to know what an MC meter is?
Why should one know how an instrument works? To understand what a measurement means.
1. Simpson 260 or 270.
These are fundamentally DC meters based on the design principle of Weston and d'Arsonval (around 1880 ) movement. This concept was a moving coil in a constant uniform magnetic field with a linear spring restrainer. Thus, a constant magnetic field in which a variable magnetic field was generated and that force restrained by a linear spring. This was not working with the Earth's magnetic field as previous meters had. Rather a locally generated magnetic was used.
A meter of this type, compared to an oscilloscope, has mass in the mechanism, and a spring, thus a mechanical resonant frequency. There is also mechanical damping so not much mechanical oscillation occurs.
I used two sine wave frequencies 14 Hz and 100 Hz, and tested a Simpson 270 ( same as 260 except somewhat greater accuracy ) and a Fluke 27. The 14 Hz was selected because at this frequency there was no DC needle fluter on the Simpson. 100 Hz is high enough that there is not much error from the input series impedance of the input coupling capacitor. Sine wave peak voltage was set at 5 V.
Note that for a full wave rectified sine wave the ratio of RMS to average is 0.707 / 0.636 = 1.112 . The Simpson was on its 10 V range, and the Fluke on auto ranging. The Simpson in AC mode passes thru any DC component. With the Simpson you need to use Output mode to remove any DC component.
Simpson 270, 14 Hz
DC mode --- No needle motion
AC mode --- 3.8 V
AC Output __ 1.6 V
Simpson 270, 100 Hz
DC mode --- no needle motion ( with none at 14 Hz there should not be any at 100 Hz )
AC mode --- 3.8 V
AC Output --- 3.6 V
Note: 5 * 0.707 = 3.54 V, and 0.707 / 0.636 = 1.112, and that 3.54 * 1.112 = 3.94 . Thus, pretty good correlation with 3.8 .
Fluke 27, 14 Hz
DC mode --- Jumps around 0.007 thru 0.043
AC mode --- 3.94 V
Fluke 27, 100 Hz
DC mode --- 0.003 V, this is probably a DC component from the Fluke function generator, could not balance it any better
AC mode --- 3.94 V
This is ball for now.
.