Drain pumps in exam rooms in obgyn

Status
Not open for further replies.

Purr24

Member
Location
Pennsylvania
Occupation
Electrician
Does a drain pump GFCI receptacle need to be wired with HCF cable in exam rooms? This receptacle is in a cabinet under the sink.
 
IMO it does.
See the definition of "Patient Care Vicinity" and "Patient Bed Location." (take note of Procedure Table) Even though the receptacle may be under a cabinet the equipment that may come in contact with the patient doesn't care. IOWs, 517.13 (A) & (B) are applicable.
 