I'm do a ground test in patient care areas, (patient vicinity)between a reference point and any exposed conductive sufaces and receptacles grounds.I'm using a fluke 87 trying to reach a resolution of 20%. Less 20 m Vrs + or - 4 mV into 1000 ohms and less than 0.1 ohms + OR - 0.02 ohms.
The readings I'm getting are between 16.0 and 19.0 on millivolt ac and .01 to .03 on resistance
leads are factored out grounds are tight, reference point is a cold water pipe and panels are bonded together as per NEC 517-14 thinking a more sensitive meter may do a better job any professional input would help. thank you dcv
The readings I'm getting are between 16.0 and 19.0 on millivolt ac and .01 to .03 on resistance
leads are factored out grounds are tight, reference point is a cold water pipe and panels are bonded together as per NEC 517-14 thinking a more sensitive meter may do a better job any professional input would help. thank you dcv