It doesn't comply with the redundant ground aspect, but I'm referring to this section:
517.13 Grounding of Receptacles and Fixed Electrical Equipment in Patient Care Areas
Wiring in patient care areas shall comply with 517.13(A) and (B).
(A) Wiring Methods. All branch circuits serving patient care areas shall be provided with an effective ground-fault current path by installation in a metal raceway system, or a cable having a metallic armor or sheath assembly. The metal raceway system, or metallic cable armor, or sheath assembly shall itself qualify as an equipment grounding conductor in accordance with 250.118.
On the spec sheet, it says:
Grounding One grounding means - Combinedarmor and full-sized aluminumground/bond conductor NEC®250.118(10)(b)
Looks to me like the MC cable complies with 517.13 (A). Do you agree?