As per NEC Article 517.60 (A)(6) FPN NO.1: It is desirable to limit the size of the isolation transformer to 10kVA or less.
As per IEC 60364-7-710:2002 Section 512.6 the rated output of transformer shall not be less than 0.5kVA and shall not exceed 10kVA.
Both NEC and IEC limit the size up to 10kVA only while the renowned supplier BENDER, HSI, ABB etc are manufacturing the rating of IPS Greator than 10kVA say 15kVA. How they are still code compliant?
Another question, Can we feed Multiple rooms say ICUs etc from a single IPS of rating 10kVA (ICUs load are approximately 2.5 to3kVA ) and we can feed 3 or 4 ICUs. Will there be any code restriction regarding multiple feed from a single IPS?
Regards
Sent from my iPhone using Tapatalk
As per IEC 60364-7-710:2002 Section 512.6 the rated output of transformer shall not be less than 0.5kVA and shall not exceed 10kVA.
Both NEC and IEC limit the size up to 10kVA only while the renowned supplier BENDER, HSI, ABB etc are manufacturing the rating of IPS Greator than 10kVA say 15kVA. How they are still code compliant?
Another question, Can we feed Multiple rooms say ICUs etc from a single IPS of rating 10kVA (ICUs load are approximately 2.5 to3kVA ) and we can feed 3 or 4 ICUs. Will there be any code restriction regarding multiple feed from a single IPS?