IEC 61010 definition of "PROTECTIVE CONDUCTOR TERMINAL"

Status
Not open for further replies.

LMAO

Senior Member
Location
Texas
I want to make sure I understand CE definition of "PROTECTIVE CONDUCTOR TERMINAL" denoted by this:

1201_OnYourMark_5019.png


IEC 61010 defines it as terminal which is bonded to conductive parts of equipment for safety purposes and is intended to be connected to an external protective earthing system. So if I ground the neutral of transformer secondary, the earth terminal is a "PROTECTIVE CONDUCTOR TERMINAL"? But if an equipment case is grounded to a terminal, the terminal is not "PROTECTIVE CONDUCTOR TERMINAL"...

Do I have this right?

thanks,
 

Sahib

Senior Member
Location
India
Protective conductor of IEC is same as EGC of NEC unless the former is ' combined one' or ' direct from earth'.
 

Tony S

Senior Member
earth%20connection_zpsmckyfwao.jpg
 

Julius Right

Senior Member
Occupation
Electrical Engineer Power Station Physical Design Retired
In my opinion, at first, this standard does not refer to power supply but:
“[Safety requirements] for electrical equipment for measurement, control, and laboratory use”. If a power supply transformer is included in this equipment envelope then its neutral could be connected with a grounding [inner] terminal.
However, still in my opinion, the protective conductor terminal has to be on the outer side of the envelope in order to facilitate " to be connected to an external protective earthing system"
 

romex jockey

Senior Member
Location
Vermont
Occupation
electrician
I would imagine just how 'protective' it actually is would be dependent on what type style earthing system one is dealing with......


~RJ~
 