I'm working up instructions for an MRI installation. The package from the MRI equipment supplier calls for the installation of an isolation transformer for convenience outlets, lighting, etc. View attachment 15723
The guidance states "THE ELECTRICAL GROUND FOR ANY AUXILIARY EQUIPMENT USED BY THE OWNER MUST BE ISOLATED FROM THE ELECTRICAL GROUND FOR THE MANUFACTURER'S EQUIPMENT. CONNECTION OF THE GROUND TO BUILDING STEEL, FOR EXAMPLE, TO THE MANUFACTURER'S EQUIPMENT GROUND IS NOT PERMITTED AND WILL CAUSE A GROUND LOOP. IF A GROUND LOOP DOES EXIST THE ELECTRICAL CONTRACTOR WILL HAVE TO INSTALL AN ISOLATION TRANSFORMER AS DESCRIBED HERE."
The sketch (hopefully attached) shows what appears to be equipment grounding conductors (bonding conductors) connecting filter enclosures, RF shield, and a power distribution panel enclosure on the secondary side of the isolation transformer to the enclosure of the isolation transformer. It does not show a connection of the isolation transformer enclosure to the building grounding electrode system. Only 120V single phase loads are supplied by the isolation transformer.
Question 1: Is the secondary neutral required to be connected to the building grounding electrode system?
Question 2: Is the isolation transformer enclosure required to be connected to the building grounding electrode system? Question 3: Is the secondary neutral required to be bonded to the isolation transformer enclosure?
Question 4: How does an isolation transformer remedy a ground loop? I would think the RF shield around the room is effectively bonded to building steel (and therefore to the building grounding electrode system), so if an equipment grounding conductor connects the RF shield around the room to the isolation transformer enclosure that would pretty much connect all the 'grounds'. Phase and neutral currents may be isolated by the isolation transformer, but the grounds?
I would like clear advice as to what to tell the electricians to do regarding connecting 'grounds'.
The guidance states "THE ELECTRICAL GROUND FOR ANY AUXILIARY EQUIPMENT USED BY THE OWNER MUST BE ISOLATED FROM THE ELECTRICAL GROUND FOR THE MANUFACTURER'S EQUIPMENT. CONNECTION OF THE GROUND TO BUILDING STEEL, FOR EXAMPLE, TO THE MANUFACTURER'S EQUIPMENT GROUND IS NOT PERMITTED AND WILL CAUSE A GROUND LOOP. IF A GROUND LOOP DOES EXIST THE ELECTRICAL CONTRACTOR WILL HAVE TO INSTALL AN ISOLATION TRANSFORMER AS DESCRIBED HERE."
The sketch (hopefully attached) shows what appears to be equipment grounding conductors (bonding conductors) connecting filter enclosures, RF shield, and a power distribution panel enclosure on the secondary side of the isolation transformer to the enclosure of the isolation transformer. It does not show a connection of the isolation transformer enclosure to the building grounding electrode system. Only 120V single phase loads are supplied by the isolation transformer.
Question 1: Is the secondary neutral required to be connected to the building grounding electrode system?
Question 2: Is the isolation transformer enclosure required to be connected to the building grounding electrode system? Question 3: Is the secondary neutral required to be bonded to the isolation transformer enclosure?
Question 4: How does an isolation transformer remedy a ground loop? I would think the RF shield around the room is effectively bonded to building steel (and therefore to the building grounding electrode system), so if an equipment grounding conductor connects the RF shield around the room to the isolation transformer enclosure that would pretty much connect all the 'grounds'. Phase and neutral currents may be isolated by the isolation transformer, but the grounds?
I would like clear advice as to what to tell the electricians to do regarding connecting 'grounds'.