medical cable duct

Status
Not open for further replies.

woodduder

Senior Member
Location
West Central FL.
In the past, I have seen drawings from certain radiology equipment manufacturers showing the 5 foot lengths of wall duct (that I would install for them and holds their cables), to be bonded at each section. If this does not appear on a specific set of drawings I have, is it required in some code? I could bond at one location and all the other pieces are connected via a manufactured coupling that is tapped and screwed by at least four screws.
 

roger

Moderator
Staff member
Location
Fl
Occupation
Retired Electrician
If the particular vendors plans do not show anything then the NEC and ACHA would be the only other parties in the game, so screws, bolts, and factory fittings would be sufficient.

Roger
 