I have an oven that is mobile for employees to use, the manufacturer labeling says to install a 50 disconnect within sight. This oven does not fall under article 430 were it is stated that a motor shall have a disconnect means within sight.
Ovens fall under article 210 to calculate kvwatt ratings per branch circuit. However, this particular oven is in a hospital, in the prostetic's room. There are no patients on this floor, and employees are the only ones permitted in this section. The question is, do I need to install a disconnect for an oven with a 50 amp cord cap that plugs into a receptacle with the disconnect is the breaker in the panel. Article 422 says that if the cap is plugged into the wall and a person has no access to it, then the disconnect shall be the main breaker.
Ovens fall under article 210 to calculate kvwatt ratings per branch circuit. However, this particular oven is in a hospital, in the prostetic's room. There are no patients on this floor, and employees are the only ones permitted in this section. The question is, do I need to install a disconnect for an oven with a 50 amp cord cap that plugs into a receptacle with the disconnect is the breaker in the panel. Article 422 says that if the cap is plugged into the wall and a person has no access to it, then the disconnect shall be the main breaker.