This is in reference to Article 517.19 (C) " Patient Vicinity Grounding and Bonding - Optional"......FPN
In the code, it is permitted to have patient equipment grounding point. The installation of said EG is permitted to be centrically or looped as convenient.
Now, in the the fine print note, it states that you are to keep the distace between the reference grounding point and patient vicinity as short as possible.
We would like to loop this EG between about 10 patient rooms. It doesnt give a specific length or distance. How are we to interpret this?
I have looked in the 2005 handbook and all examples show only one patient room. I would love to know the exact meaning of this.Thanks for any help.