Anybody ever work on one of these? Its an Olympus AU2700 medical analyzer for blood, urine etc...
I was told L6-30R receptacles but when I got to the job I found out they are using a UPS.
The spec sheet calls for 50 amp circuit breaker when used with UPS.
I'm trying to figure out if I'm running #10 on a 50 amp breaker with a 30 amp receptacle or
#8 on a 50 amp breaker with a 30 amp receptacle.
The company hasn't been much help yet.
The unit isn't on site and the circuit needs to be ready to go as soon as it arrives.