Patient Care Space

Status
Not open for further replies.

Coachmike

Senior Member
Location
Morris County, NJ
Occupation
Electrical Contractor
Definition of “patient care space”?
Does the nurses station, Room 129, in this recovery room require the mechanical protection referenced in 517.C.3?
 

Attachments

  • D5FBF775-A05C-4482-AFFC-23F74145D194.jpeg
    D5FBF775-A05C-4482-AFFC-23F74145D194.jpeg
    1.2 MB · Views: 7
Part III which is where 517.30(C)(3) is found and is for all areas. Part II is specific for Patient Care Areas so the answer is yes they do have to have the mechanical protection as required in 517.30(C)(3)

Roger
 