Power Requirements

Status
Not open for further replies.

mshields

Senior Member
Location
Boston, MA
A piece of medical equipment I am to provide power to has the following Power Requirements

* 1N PE 230VAC, 50/60HZ, 10kVA

What does the 1N PE part signify?
 
I wonder if it is just an abbreviation for "input voltage".

IN instead of 1N, and PE is an abbreviation for "potential" or voltage.
 
I would expect it to be:
1 neutral, &
Potential Earth (ground), &
230 volt conductor

The diagram does show this configuration; but since it shows the neutral also switched, I would suspect it would work fine on a 2 pole circuit of your 208 or 240. I would also expect a rep to say it wouldn't be acceptable.
 
Last edited:
This is an IEC terminology system, not meant for North America. It means:

1 = single (1) phase conductor. For them, 2 "hot" conductors out of a 3 phase distribution scheme is called "2 phase", which has a totally different meaning here.

N = Neutral, meaning the current return path for a grounded secondary of a distribution system

PE = Protected Earth, our equivalent of an EGC

So understand that basically in most other countries, there is no single phase distribution system, everyone gets 3phase, even if they only USE it as single phase. In general, the "harmonized" official voltage is 400Y230V, so people get 400V phase to phase, 230V phase to Neutral. So in this piece of equipment they are telling the installer to connect only 1 phase + the Neutral to get 230V, then a separate equipment grounding conductor (protected just means an insulated grounding conductor).

Our 240V distribution system however does not work that way, single phase for us is 240V line to line, not line to neutral. If they are not using internal ground connections for components inside of the unit, and they should not be, then all that really matters is that you give it the 240V. But do NOT give it one hot + N from our distribution system, because that will be only 120V.

However, if under8ed is correct and it can accept anything from 100-300V, then even 120V should work, but the current draw would be twice what it would be at 240V. At 10kVA, that's a lot of 120V current.
 
I see now that the PDF link is not from the OP, so unless that did happen to be your piece of equipment; you should ignore the wide voltage range in my post #5. We do sometimes hook up European equipment to our voltages using two hots; but this is done either with knowledge of how the conductor intended as a neutral is used, or we have verified the installation with the manufacturer.
 