joedanzhere
New User
- Location
- New Jersey
- Occupation
- Hospital Electrician
I work as an electrician in a hospital, and I don't quite understand Line Isolation Monitor panels. I have two legs which read 60 volts each, which means there is no neutral or (grounded conductor). I understand this part. I have heard some people say that you should not be able to see any voltage between either leg and the ground. In the circuits I've seen from the LIM panels I get a reading of 60 volts from either leg to ground. From what I understand the grounding conductor is bonded to the enclosure of the secondary of the LIM transformer. So if I am supposed to get a reading of 0 volts to ground on my receptacles this would mean that my LIM transformer secondary enclosure is not bonded to my mains electrical system. Since I am getting a reading of 60 volts to ground on my LIM system does this mean that my system is not grounded properly? Also from my understanding if you have a ground fault in a LIM system where a hot wire touches a metal enclosure nothing should trip but just alarm my LIM monitor. If I am read 60 volts to ground this would lead me to believe that my ground is bonded to my mains electrical system and that if I had a fault where one hot wire touched a metal enclosure I would trip a breaker, which from what I understand should not happen. Basically my question is should I be able to see a potential of 60 volts from either leg to ground in a LIM system and is this because my system is bonded somewhere where it shouldn't be?