carl.arnsdorff
We manufacture medical scanning equipment.
The equipment has an option for a Power Distribution Unit (PDU) that can feed several different configurations of scanners which also includes adjustable circuit breakers and isolation transformers, etc.
The main circuit breaker for the PDU is a 250 amp breaker. Is it a code violation to specify the feeder circuit breaker from the building supply to have a lower ampacity than the PDU (e.g. 175 amp)?
The end equipment after the PDU does not require the higher amperage and the equipment itself has circuit breakers rated for the load it will pull.
I am aware that the industrial equipment section does not allow a smaller breaker feeding a larger breaker but I'm not thinking this is applicable here. Can anyone hlep me with this?