A company is installing a xray unit in a doctors office. The xray unit has a a part that moves. This part has 2 SO cables coming out of it. The xray installer ran these two cables through a hole in a cover plate on a wall box. In the wall box, he butt spliced the SO cord to THHN wire. The THHN runs in conduit and connects to the xray unit.
The local inspector says SO cord can't be used as a substitute for fixed wiring. He wants attachment plugs placed on the SO cords. The THHN would then terminate on receptacles (nonstandard). (This isn't normal 120V power, its part of the Xray unit circuitry).
I think 400.7 (A) (10) allows the SO cable for connection of moving parts, and 400.14 allows the cable to run through a cover with a bushing. I am also worried that we are adding plugs to the manufacturer's equipment design (ie, we are modifying the equipment).
The inspector doesn't buy my argument. Please give me your opinion. Do you think the receptacles and plugs are required, or not?
Steve
[ October 24, 2003, 05:26 PM: Message edited by: steve66 ]
The local inspector says SO cord can't be used as a substitute for fixed wiring. He wants attachment plugs placed on the SO cords. The THHN would then terminate on receptacles (nonstandard). (This isn't normal 120V power, its part of the Xray unit circuitry).
I think 400.7 (A) (10) allows the SO cable for connection of moving parts, and 400.14 allows the cable to run through a cover with a bushing. I am also worried that we are adding plugs to the manufacturer's equipment design (ie, we are modifying the equipment).
The inspector doesn't buy my argument. Please give me your opinion. Do you think the receptacles and plugs are required, or not?