How is this to be applied in the case there is a feeder to a panel (blue in image below), that then has a branch circuit (green in image below)to a single x-ray unit. The feeder article indicates 'two or more'... 'not less than 50%....'. But, leaves opaque what happens if a single unit.
Is one to interpret that since 51.7.73(A)(1) states 50 of momentary or 100% of long-time (whichever greater), that for the feeder a similar comparison must be made between the momentary vs. long-time current for the feeder (blue)?
Is one to interpret that since 51.7.73(A)(1) states 50 of momentary or 100% of long-time (whichever greater), that for the feeder a similar comparison must be made between the momentary vs. long-time current for the feeder (blue)?