Found where conductors were run in PVC conduit underground to patient care areas. 517.13 A requires conductors to be installed in a conduit that is approved as a equipment grounding path as noted in 250.118. In leu of the fact that the clinic is open for business, would the idea of installing a #6 ground wire from the panel and tapping off of each of the conduits with a #10 wire and a pipe clamp be acceptable.
Thanks in advance for your answer.
Ed