Hello.
I have a situation where i have an electrical panel in a hospital room. This electrical panel is staying and a medical x-ray machine table is being added in-front of the panel. The length of the table will be about 2'6" away from the front of the electrical panel. This table is adjustable type where it can move farther to one side where we can have a clearance of 3' feet temporarily (if works need to be done on the panel). Kind like of having a electrical panel in a closet with the doors closed and when open the doors the panel meets the clearance.
I don't see violation of the panel clearance, as long as we can provide 3feet clearance in-front of it temporarily. Is there an issue that i'm not seeing?
Thank you
I have a situation where i have an electrical panel in a hospital room. This electrical panel is staying and a medical x-ray machine table is being added in-front of the panel. The length of the table will be about 2'6" away from the front of the electrical panel. This table is adjustable type where it can move farther to one side where we can have a clearance of 3' feet temporarily (if works need to be done on the panel). Kind like of having a electrical panel in a closet with the doors closed and when open the doors the panel meets the clearance.