charlie tuna said:i see the necessity for their use mainly in the medical field where absolute grounding is necessary to remove even static charges outside the electrical system's wiring.... where they terminate in their own ground system independent of any other ground..
Charlie, in medical applications quite the opposite is true, there are very few if any IG's and the ones that are installed are probably in violation or a mistake.
Patient rooms must have an Equipotential of <.1 ohm and this is hard to achieve with isolated grounds and numerous paths.
Most medical equipment manufacturers of equipment like C.T.'s, Xray's, Linear Accelerators, etc... have multiple bonding paths that are connected in parallel by design.
Roger