gar
- Ann Arbor, Michigan
- EE
090726-2059 EST
In 63 or 64 I bought a Ballantine 420 AC and DC Calibrator. This used an OA2 voltage regulator tube for the reference. Clearly these were aged and selected for stability.
For convenience in May 1964 I built a small DC voltage reference box with outputs of 12.1, 10.0, 1.0, and 0.1 V. This was a voltage divider from a nominal 12.1 V temperature compensated Zener diode. This box was calibrated to the Ballantine at 10.0 V using a voltage difference method.
Today my voltage reference reads the following using a Beckman 4410 meter:
Input DC
AC V Ouput
100 --- 9.976
110 --- 9.978
120 --- 9.980
130 --- 9.982
100 --- 9.978
130 --- 9.982
15 minutes later
120 --- 9.981
At 120 V input to both the Ballantine and my box the reading of my box was
0.0054 V higher than the Ballantine.
Rather good stability of the two devices over the years.
These readings do not give me an absolute value for either box. A Fluke 27, lower resolution than the Beckman, produced a reading of 9.98 V. I would conclude that the Ballantine was slightly off calibration when new and still is. Did the Ballantine drift 0.0054 over the years, or my Zener, or both. No way to know. It is more surprising that the OA2 was this stable over this period of time. However, the Zener diode was from the early days of making these devices.
Input DC
AC V Ouput
100 --- 9.976
110 --- 9.978
120 --- 9.980
130 --- 9.982
100 --- 9.978
130 --- 9.982
15 minutes later
120 --- 9.981