An existing, diagnostic xray machine is being relocated from one clinic to another. I have not worked on a xray machine this big before.
I will need to install a separate switch in the service switchgear (480V). I will also install a 50 KVA transformer in the xray room (no other place to put it). It will be a 1 phase 240V w/ neutral secondary. The neutral will not be used with the xray machine.
The inspector is requiring me to verify the service can handle the new load of the xray machine. I'm confused as to exactly what that load will be.
The specs on the xray generator are:
maximum KW 37.5
KW @ 100KV @ 0.1 sec 30
Momentary line current 345 amps
The xray tech told me the "resting load" would be around 20 amps.
What do I use to determine the load this xray machine will add to the service.