Those patients have afib and the DC is in measured pulses of around 5ms. That is totally different from the effects of steady current DC on a healthy person.
Mark, K8MHZ
To add to the measured levels.
In the 70's, I worked at a university hospital, where the JHAC required testing for all electrical devices
which might be in contact with patients. The main concern was stray currents into arteries and veins.
The device I used was a electronic Amp-Meter which registered clearly down to 10 microAmps.
The voltage was not the main issue, only the current flowing from a medical device which could exceed 10 microAmps.
In my research time there,
I commonly measured human skin-surface resistance, point to point, at 10 KiloOhms and upwards to 1 MegOhm.
Commonly, the patient skin-resistance was found to be 50 KiloOhms.
Given that scenario, the Calculated applied Voltage could be 0.5 Volts.
That is 0.5 Volts applied in a laboratory experiment.
Given that some leakage could be via in-the-vein connection, the resistance of the system drops to the 10 KiloOhm range.
(1) D.C. could cause heart contractions, similar to a defibrillator, if momentary.
(2) A.C. could cause heart fibrilations, known as multiple-nodes-of-contraction,
wherein the atria will show chaotic depolarisation with multiple foci.
Visually, this appears to be a dozen points of contraction, instead of one single contraction for a heart-beat.
It has been a very long time passing, since those experiences.
But, it should be obvious that testing for electrical leakage
flowing from devices/equipment
is a serious concern.
This is a good thread. Thank you for the contribution, Mark.