Can anyone please tell me if there is any special electrical requirements for Diagnostic Imaging (DI) rooms for outlets installed on the Lead shielding wall? Are there any special requirements for the receptacles, jb's, conduits, etc?
Thanks!
 

roger

Moderator
Staff member
Location
Fl
Occupation
Retired Electrician
Can anyone please tell me if there is any special electrical requirements for Diagnostic Imaging (DI) rooms for outlets installed on the Lead shielding wall? Are there any special requirements for the receptacles, jb's, conduits, etc?
Thanks!

They just have to be within the lead envelope, this means you or someone may have to add lead behind the boxes if the installation penetrates the envelope this would include screws. Other than that, there is nothing special as far as the raceways and devices themselves