Does x-ray machine in hospital require a redundant ground?

Cartoon1

Senior Member
Location
Florida
Occupation
Electrical Engineer
I have an X-ray machine that has 80amp fla. It also requires the the ground to be sized the same as the hot wires. Requires 3 hot wires and no netrual. 3 phase 480v.

I'm providing a 100A breaker with 3#2 and 1#2 green ground. All copper.

My question is does the X-ray require a redundant ground cable? And if so, what size?

Thank you
 

roger

Moderator
Staff member
Location
Fl
Occupation
Retired Electrician
I have an X-ray machine that has 80amp fla. It also requires the the ground to be sized the same as the hot wires. Requires 3 hot wires and no netrual. 3 phase 480v.

I'm providing a 100A breaker with 3#2 and 1#2 green ground. All copper.

My question is does the X-ray require a redundant ground cable? And if so, what size?

Thank you
As Larry says, it's there by default and the #2 is it. I assume your designing per the manufacturers plans and many times the equipment and room is listed as a unit.
 