I am working on a project to automate fuse testing. Per the test procedure, thermocouples are attached directly to the fuse furrules. A current is passed through the fuse until thermal stability is achieved. The TCs are going to be connected to an amplifier then to an A/D converter and finally to a microcontroller for analysis.
Here’s the issues I am having that I do not know how to solve.
The power supply we use to drive the current will drive it at whatever voltage is conveient to achieve the curent required. So, if the fuse should open, there is the possiblity of imposing severeal hundred volts onto the thermocouple for a brief time. I say brief, but the few milliseconds will be sufficient to damage/destroy the electronics downstream of the TCs. What I need is a way of electrically islotaing the TCs while permitting the 0-15mV signal from tem to reach the amplifier undisturbed.
I do not want to mechanically isolate the TCs from the fuse as the introduction of any medium between the fuse and TC will introduce errors in temperature measurement. While that is not necessarily a chritical attribute, it will make calibration traceable to NIST more difficult.
Do any of you brilliant guys out there know of a solution that is not off the shelf technology (in other words I need discrete component type stuff) or bulky to the point of needing it’s own desk (I need it all to fit in a box the size of a deck of cads)?