080721-0956 EST
To those that have Gauss meters:
1. Why do you have one of these meters?
2. What is your use?
3. What is the resolution (smallest increment)?
4. Manufacturer and model number?
5. What type of sensor?
6. How large is the sensor head?
7. How often is it used?
8. Do you have customers that require these measurements?
9. Do you also have an electric field intensity meter?
I have a coil that I bought many years ago that is calibrated at 1 gauss @ 60 Hz = 20 MV. With my Fluke 27 this gives me a resolution of 5 mG. I have added a 200 gain amplifier thus improving the resolution to 0.025 mG. Currently my noise level is about 6 times this, or 0.15 mG. That should be correctable with filtering.
Some of my observations:
On the ground below my 3 phase primary lines I read about 0.05 mG above the noise level.
The ground current to my water pipe is about 110 MA. Using the probe it appears to flow both into the house and earth. Maybe equally or not, but it looks equal at this time. I do not have a large current probe to go around the water pipe so I can not perform a correlation.
All my water plumbing is copper, all vertical drains in the basement are cast iron. It is interesting that the magnetic field probe sees current in one of the drains in the basement. My water line runs under the basement floor, but with my present sensitivity and noise level I can not detect its location from the magnetic field probe and the low 55 MA current.
This is a coil type probe and therefore when not moving is insensitive to steady magnetic fields. It is interesting that when I move the probe I get a short time rise in the reading. This of course results from moving the coil in the earth's magnetic field.
.
To those that have Gauss meters:
1. Why do you have one of these meters?
2. What is your use?
3. What is the resolution (smallest increment)?
4. Manufacturer and model number?
5. What type of sensor?
6. How large is the sensor head?
7. How often is it used?
8. Do you have customers that require these measurements?
9. Do you also have an electric field intensity meter?
I have a coil that I bought many years ago that is calibrated at 1 gauss @ 60 Hz = 20 MV. With my Fluke 27 this gives me a resolution of 5 mG. I have added a 200 gain amplifier thus improving the resolution to 0.025 mG. Currently my noise level is about 6 times this, or 0.15 mG. That should be correctable with filtering.
Some of my observations:
On the ground below my 3 phase primary lines I read about 0.05 mG above the noise level.
The ground current to my water pipe is about 110 MA. Using the probe it appears to flow both into the house and earth. Maybe equally or not, but it looks equal at this time. I do not have a large current probe to go around the water pipe so I can not perform a correlation.
All my water plumbing is copper, all vertical drains in the basement are cast iron. It is interesting that the magnetic field probe sees current in one of the drains in the basement. My water line runs under the basement floor, but with my present sensitivity and noise level I can not detect its location from the magnetic field probe and the low 55 MA current.
This is a coil type probe and therefore when not moving is insensitive to steady magnetic fields. It is interesting that when I move the probe I get a short time rise in the reading. This of course results from moving the coil in the earth's magnetic field.
.