There is one dedicated circuit required for what packet calls “ICP-OES”Just to reiterate what Augie said, for an individual branch circuit, in conditioned spaces, neither the 80% limit nor the temperature correction apply. It's as simple as "non-motor load <= 15A; use a 15A breaker with #14 wire."
Cheers, Wayne
P.S. In the OP you have highlighted the lines for both the spectrometer and the water chiller. These are each getting separate circuits? Does either one use a motor?
However engineer recently asked me to install two dedicated circuits