Hospital Grade receptacle

Status
Not open for further replies.
If it is mentioned in some medical equipments that proper grounding shall be provided. is it meaning that we should provide a hospital grade receptacles for that device OR we shall provide a separate grounding systems for medical equipments/areas which we already going to do.

What if we provide this separate grounding systems for medical equipments/areas and going to use normal receptacle there is any impact on the patient life safety?:roll:

In brief what is the difference between hospital grade and normal grade receptacles from the grounding point of view?
 

JJWalecka

Senior Member
Location
New England
If it is mentioned in some medical equipments that proper grounding shall be provided. is it meaning that we should provide a hospital grade receptacles for that device OR we shall provide a separate grounding systems for medical equipments/areas which we already going to do.

What if we provide this separate grounding systems for medical equipments/areas and going to use normal receptacle there is any impact on the patient life safety?:roll:

In brief what is the difference between hospital grade and normal grade receptacles from the grounding point of view?

A hospital grade receptical has greater contact tension. And isnt always required.
the equipotential plane is the more important issue in most cases.
517.13 requires a redundant grounding path be established by using a separate insulated equipment grounding conductor in conjunction with a totally separate grounding path on either a metal raceway or a cable with an outer metal armor or metal sheath identified as an acceptable grounding return path.
 