Panel MCB sizing for (2) CT machines and (2) Xray machines

CXC0987

Member
Location
NYC
Occupation
ELECTRICAL ENGINEER
Sizing a panel for (2) CT machines and (2) Xray machines, NEC 660.6 states that the ampacity of supply branch-circuit conductors and the overcurrent protective devices shall not be less than 50 percent of the momentary rating or 100 percent of the long-time rating, whichever is greater.

this apply to X-ray machine but does that also apply to CT scanning machine? I don't there is any code section in NEC or NFPA 99 talking about CR machine demand power calculation.
 

LarryFine

Master Electrician Electric Contractor Richmond VA
Location
Henrico County, VA
Occupation
Electrical Contractor
Welcome to the forum.

Do CT scanners have similar momentary/long-time operating characteristics?

If not, then use the long-time current like you would with most loads.
 

CXC0987

Member
Location
NYC
Occupation
ELECTRICAL ENGINEER
Hi Larry, thanks for the response, appreciated
the CT almost has 3 times the momentary KVA than the xray.
I am dealing with Siemens CT machines, and their site-specific gives you a chart showing the KVA consumption when CT operates at 2 sec, 35 sec, and 100 sec (350kva, 130kva, and 87kva).
Not too sure if I should use KVA rating at 35sec or 100sec which really depends on the definition of long-time current.
 