Is the QO120MH acceptable replacement for a slow action fuse?
Sent from my SM-G930V using Tapatalk
Sent from my SM-G930V using Tapatalk
Is the QO120MH acceptable replacement for a slow action fuse?
Sent from my SM-G930V using Tapatalk
Open a photobucket account, upload your photo. When viewing an individual photo there will be options for sharing, one of them is IMG code. click on that and it will copy the code to your clipboard, and in the place you want to place that image within your post, paste the IMG code. It will display the code when you are composing your post but after you submit it that code gets replaced with the image when anyone is viewing the post. May be some exceptions if someone has firewall settings or browser settings blocking it, but will show for most people. If you remove the image from photobucket, then the IMG code will be pointing to an invalid address and there will be no image shown.I downloaded photo bucket. How do I post a link for a photo? I would like someone to look at the power requirements and give an opinion.
Sent from my SM-G930V using Tapatalk
I think I still need to install a fused disconnect or get approval from Henry Shein for a QO120MH breaker instead if fused disconnect.Open a photobucket account, upload your photo. When viewing an individual photo there will be options for sharing, one of them is IMG code. click on that and it will copy the code to your clipboard, and in the place you want to place that image within your post, paste the IMG code. It will display the code when you are composing your post but after you submit it that code gets replaced with the image when anyone is viewing the post. May be some exceptions if someone has firewall settings or browser settings blocking it, but will show for most people. If you remove the image from photobucket, then the IMG code will be pointing to an invalid address and there will be no image shown.
I also agree that you probably want to use a QO120HM breaker.
If it calls for a fuse, you about need to install a fused disconnect, somewhere. Or find an old FSP Loadcenter:blink:I think I still need to install a fused disconnect or get approval from Henry Shein for a QO120MH breaker instead if fused disconnect.
The fact it doesn't state fuse/ breaker and only says "fused at the distribution panel, means I have to be fused, wouldn't this be correct?
Wait a second, how do I become fused at the distribution panel when it is a QO panel? Any thoughts. Out of 5 electrical inspectors, this specific one is the worst to deal with. He is always an A-hole.
Sent from my SM-G930V using Tapatalk
Is 21.3 the momentary rating or the long time rating? If long term rating I think you do need 10AWG. Otherwise the greater value of:100% of long term rating or 50% of short term rating.I think I'm going to have to go to a fuse disconnect.
I'll find out Monday if the inspector will approve.
He originally wanted 10awg ran since the nameplate said 21.3 amps. Article 517.73a1 states momentary sized to 50%. He didn't agree.
Sent from my SM-G930V using Tapatalk
The equipment literature said the maximum current rating is momentary.Is 21.3 the momentary rating or the long time rating? If long term rating I think you do need 10AWG. Otherwise the greater value of:100% of long term rating or 50% of short term rating.
I'm not a medical equipment expert, but dental X-ray machines may not even have a long term rating. Seems that would apply more to CT scanners.
What fuse rating and type does it call for?The equipment literature said the maximum current rating is momentary.
The inspector just saw a nameplate that said 21.3amps. What I'm worried about is it also says fused at the distribution panel. That is my only concern now, It can not be accomplished with a single phase QO panel.
I'll try to present this information to the inspector and see what he says. That's all I can do. At least I can say it is not.my fault and I tried to do what I could.
Sent from my SM-G930V using Tapatalk
I just posted a picture. So hopefully it worked. The column EX-1 is what I am dealing with.What fuse rating and type does it call for?
This is by far the worst experience with medical supply company I have had. The plans originally stated a 240vac, 20amp outlet and then came back and said that the Canada and US version is only 120vac.AFAIK "slow" isn't normally a used as fuse type designation, they probably do want some time delay characteristic, but still left it wide open with "slow".
Plug fuses - S and T are fairly common, cartridge fuses - they should specify RK1, RK5, J....
I think more verification from the equipment provider is necessary
Let me guess, made in an Asian country?This is by far the worst experience with medical supply company I have had. The plans originally stated a 240vac, 20amp outlet and then came back and said that the Canada and US version is only 120vac.
I will have to call Monday to get approval to provide for inspector saying we can use a breaker. I really don't have anywhere to mount a fused disconnect anyway.