gar
Senior Member
- Location
- Ann Arbor, Michigan
- Occupation
- EE
080808-0834 EST
I have found little additional information in the on-line manuals.
In a software manual it seems to imply that the maximum baud rate is 19.2 kbaud, and that 8 data bits and 1 stop bit are mandatory. Apparently Xon-Xoff is optional, and there is no option for hardware handshake. Where I found this information no mention was made of parity, but elsewhere it appears parity must be none.
Post #1 presents a problem with a disparity of readings between the 43B and the data collected on a laptop from the 43B. To try to understand the problem it is necessary to have precise knowledge of how the 43B makes measurements and how these are communicated to the laptop.
The manual does not give an accurate unambiguous description of what happens in the sag-swell measurement. In particular there is no information on what is transferred from the 43B to the RS232 port. I believe we have data every 1 second of the min, max and average of the one cycle RMS values. Is the 1 second data always output to RS232 independent of any other settings of the 43B? If so, then on the computer you should be able to look at all data down to the 1 second sample independent of how much data has been collected up to some maximum. One day is 86,400 seconds and if 100 bytes are required for the 1 second data, then about 8 megabytes are required per day. I think 100 bytes per sample point is quite high. Any modern computer will have no problem storing a number of days of this data.
The data and the readings on the 43B will change as the data collection time increases because of compression. This is because of limited data storage capacity on the 43B. Compression takes the most min and most max of the samples of the interval being compressed and these become the new min and max values and the average of the averages is calculated to become the new average. The fine time increment data is lost at the 43B.
I believe the numeric values displayed on the 43B will be the values of the sample at the cursor location. If the screen width is greater than 4 minutes, then the numeric values at a given real time may be different than the 1 second sample value at that same real time point. A real time sample point after compression is a composite of a number of 1 second sample points used to generate the compressed value. Does this have anything to do with the problem of post #1?
From post #1:
ptonsparky:
What additional information can you provide on the settings of the 43B and the computer when these conflicting results were obtained?
.
I have found little additional information in the on-line manuals.
In a software manual it seems to imply that the maximum baud rate is 19.2 kbaud, and that 8 data bits and 1 stop bit are mandatory. Apparently Xon-Xoff is optional, and there is no option for hardware handshake. Where I found this information no mention was made of parity, but elsewhere it appears parity must be none.
Post #1 presents a problem with a disparity of readings between the 43B and the data collected on a laptop from the 43B. To try to understand the problem it is necessary to have precise knowledge of how the 43B makes measurements and how these are communicated to the laptop.
The manual does not give an accurate unambiguous description of what happens in the sag-swell measurement. In particular there is no information on what is transferred from the 43B to the RS232 port. I believe we have data every 1 second of the min, max and average of the one cycle RMS values. Is the 1 second data always output to RS232 independent of any other settings of the 43B? If so, then on the computer you should be able to look at all data down to the 1 second sample independent of how much data has been collected up to some maximum. One day is 86,400 seconds and if 100 bytes are required for the 1 second data, then about 8 megabytes are required per day. I think 100 bytes per sample point is quite high. Any modern computer will have no problem storing a number of days of this data.
The data and the readings on the 43B will change as the data collection time increases because of compression. This is because of limited data storage capacity on the 43B. Compression takes the most min and most max of the samples of the interval being compressed and these become the new min and max values and the average of the averages is calculated to become the new average. The fine time increment data is lost at the 43B.
I believe the numeric values displayed on the 43B will be the values of the sample at the cursor location. If the screen width is greater than 4 minutes, then the numeric values at a given real time may be different than the 1 second sample value at that same real time point. A real time sample point after compression is a composite of a number of 1 second sample points used to generate the compressed value. Does this have anything to do with the problem of post #1?
From post #1:
Was the 43B in sag-swell mode? What was the screen width in time? Was 2.155 the maximum reading, or a peak meaning something else? Or was the meter in transient mode? I do not think that transient and sag-swell can run simultaneously.Meter display caught a 2.155 volt spike. Laptop displays a max of 1.096 volt. What gives?
ptonsparky:
What additional information can you provide on the settings of the 43B and the computer when these conflicting results were obtained?
.