76nemo:
Do you work at Sterling also?
To start over. What I think I understand about the 43B in the sag-swell mode is:
1. At a high sampling rate, maybe about 16/256 MS, the input signal is sampled and from these samples the RMS value for for one full cycle at 60 Hz is calculated. Whether it is 256 samples or 512 in one cycle is not important or whether the RMS value is calculated for 1/2 cycle or a full cycle again may not be important.
2. Important is that we get an RMS measurement every 1/60 of a second.
3. Next important point is that at the end of a 1 second collection period that we get 3 measurements from the input signal. These are used for the internal display. Next everything is reset and we start a new 1 second collection period from another set of 60 (maybe 120) measurements.
None of these 60 measurements are available for display internally or externally. Only the summary values min, max, and average are available.
4. In sag-swell mode the only available data values are the results of each 1 second period. These values are the min. max, and average of the 60 measurements of the respective 1 second period. Also collected is a time stamp.
5. Every 1 second period I would expect the 43B to be able to send these 3 values plus the real time to a collection computer. If 1200 baud is too slow for the way this data is formatted to transfer all this data, then certainly at the maximum baud rate of 19.2 kbaud the transfer can be performed. But I could certainly send that amount of data at 1200 baud. 1200 baud is about 120 characters per second.
6. For internal data there is compression after the first 240 seconds, 4 minutes, unless the buffer is cleared and the initial 4 minute period is restarted.
Fluke claims there is too much data to send min, max, and average every 1 second and therefore only sends the average value for continuous monitoring..
The too much data argument I do not buy.
.