211106-1720 EST
Further study on the Fluke 87,
This time I am using a Tektronix function generator as the signal source. Each different signal is a steady state repetitive waveform, either a sine wave or a square wave. Both have a long term average value of 0 ( no DC component ), and are symmetrical in time. I have taken the actual measured values and rounded them to their theoretical values so you can easily see the differences.
The test frequencies were 0.2 Hz, 20 Hz, and 200 Hz.
Peak voltage for all waveform was +/- 5 V.
The Fluke 87 meter settings were
DC 1 mS peak, 100 mS, and
AC 1 mS peak, 100 mS.
................... DC 1 mS peak ......... DC 100 mS ...................... AC 1 mS peak ,,,,,,,,,,, AC 100 mS
0.2 HZ
Square ...... - 5 V .....+ 5 V .......... - 5 V ....... + 5 V ................. - 10 V ...... + 10 V ........ 0 V ...... 2.08 V
Sine .......... - 5 V ..... + 5 V ......... - 5 V ....... +5 V .................. -0.8 V ...... + 0.8 V ........ 0 V ...... 0.16 V
20 HZ
Square ...... - 5 V .....+ 5 V .......... - 0.1 V ....... +0.5 V ................. - 5.5 V ...... + 5.5 V ........ 5.0 V ...... 5.0 V
Sine .......... - 5 V ..... + 5 V ......... - 0.3 V ....... +0.5 V .................. -5.0 V ...... + 5.0 V ........ 3.6 V ...... 3.6 V
200 Hz
Square ...... - 5 V .....+ 5 V .......... - 0.2 V ....... +0.2 V ................. - 5 V ...... + 5 V ........ 5.0 V ...... 5.0 V
Sine .......... - 5 V ..... + 5 V ......... - 0.0 V ....... +0.0 V .................. -5 V ...... + 5 V ........ 3.6 V ...... 3.6 V
If you saw my raw data you might disagree with my rounding, but I think it adequately describes what the meter is doing.
The results for "0.2 Hz, AC 1 mS peak, Sine wave" look out of place and uncorrelated, but I rechecked and basically saw the same results.
I haven't done the transformer inrush test yet.
.