K kentaylor04 Member Feb 27, 2005 #1 have a 240 v 1 phase machine to install. service is 120/240 delta 3 phase will the high leg of this service cause trouble for x- ray machine
have a 240 v 1 phase machine to install. service is 120/240 delta 3 phase will the high leg of this service cause trouble for x- ray machine
B bphgravity Senior Member Location Florida Feb 27, 2005 #2 Re: X-RAY MACHINE no. Just don't hook up any line-to-neutral 120V loads to the high leg phase. [ February 27, 2005, 09:21 AM: Message edited by: bphgravity ]
Re: X-RAY MACHINE no. Just don't hook up any line-to-neutral 120V loads to the high leg phase. [ February 27, 2005, 09:21 AM: Message edited by: bphgravity ]
K kentaylor04 Member Feb 27, 2005 #3 Re: X-RAY MACHINE thanks i was thinking about, was not sure so i will set a 3 pole breaker an use 2
K kiloamp7 Senior Member Feb 27, 2005 #4 Re: X-RAY MACHINE Or you can set a 2-pole bkr. if you wish, & use 2-poles.