chevyx92
Senior Member
- Location
- VA BCH, VA
I ran into a situation today where the xray machine wouldn't work, i found the breaker in the room was tripped and wouldn't reset. While troubleshooting I found there is a under voltage release accessory in the breaker. There was no under voltage condition but it wouldn't allow me to reset the breaker. This accessory operates at 480 volt. Once bypassed the breaker would reset. I determined the under voltage release to be bad. This is also tied into the emergency stop buttons in the rom . My question is, is this a common setup to have and under voltage release instead of a shunt trip?
